NDMA handling - safety measures

Hi all :slight_smile:

During the Nitrosamines and GTIs conference (Vienna, April 2025), we discussed that NDMA is a volatile nitrosamine and should therefore be handled with particular caution.
As a Finished Product (FP) manufacturer, we have typically outsourced such analytical activities. However, we are now considering initiating in-house analytical work for nitrosamines in several products, including those related to Metformin.
Are there any available publications or guidelines outlining the safety measures we should consider?

Thank you.

@elenipoliti This is a topic that has been really high in my list of concern when it comes to confirmatory testing. Countless times, I have seen nitrosamine reference materials being treated as any other reagent in the lab. We are talking about a highly potent impurity at a very high purity in the hands of the analyst.
Unless a lab is equip with the necessary infrastructure and appropriate SOP and training, analyst should not be handling solid nitrosamines reference materials.
I would like to invite @parthamuk2003 @mayank.bhanti @jason.brown to share their perspective and experience in the lab operation when it comes to safety

2 Likes

Dear @Naiffer_Host thank you for your feedback.
I would love to hear rest colleagues’ opinion on this matter.

Hi Eleni,

in few words, the best solution is a dedicated laboratory equipped with fume hoods with glove-boxes, in order to avoid any possible contact with carcinogenic substances. We choose this solution.
(I do not know if a glove box may be installed in an existing fume hood).

Regarding the guidelines, a good starting point is the:
Guidance for the safe management of hazardous medicinal products at work
It was written for the management of mutagenic/carcinogenic medicnal products in hospitals, but many concepts about CMR substances are valid in general way.

If you cannot install a glove box, the best solution is the use of commercial diluted standards; even if they are still classified as carcinogens (H350), they are less dangerous than pure NDMA.
Of course you should discuss all the details with your HSE department.
kind regards

1 Like

@elenipoliti, You may also refer to the video tutorial on handling nitrosamine reference standards. Here is the link: https://youtu.be/EYTEoDkqu8c?si=ffJ9QKJbN-65BeIw

1 Like

We are rather fortunate in that we already had a balance enclosure next to our LC-MS. So, standard preparations can occur there. This will need to be proceduralized. After that first dilution, the concern is minimal, as previously alluded to. Samples are not considered a primary concern due to the concentration within the sample.